Review of Radiation Oncology Physics: A Handbook for Teachers and Students

CHAPTER 5.

ERVIN B. PODGORSAK

Department of Medical Physics

McGill University Health Centre

Montréal, Québec, Canada

5.1. INTRODUCTION

Since the inception of radiotherapy soon after the discovery of x-rays by Roentgen in 1895, the technology of x-ray production has first been aimed toward ever higher photon and electron beam energies and intensities, and more recently toward computerization and intensity-modulated beam delivery. During the first 50 years of radiotherapy, the techno-logical progress has been relatively slow and mainly based on x-ray tubes, Van de Graaff generators and betatrons.

The invention of the cobalt-60 teletherapy unit by H.E. Johns in Canada in the early 1950s provided a tremendous boost in the quest for higher photon energies, and placed the cobalt unit into the forefront of radiotherapy for a number of years. The concurrently developed medical linear accelerators (linacs), however, soon eclipsed the cobalt unit, moved through five increasingly sophisticated generations, and became the most widely used radiation source in modern radiotherapy. With its compact and efficient design, the linac offers excellent versatility for use in radiotherapy through isocentric mounting and provides either electron or megavoltage x-ray therapy with a wide range of energies.

In addition to linacs, electron and x-ray radiotherapy is also carried out with other types of accelerators, such as betatrons and microtrons. More exotic particles, such as protons, neutrons, heavy ions, and negative π mesons, all produced by special accelerators, are also sometimes used for radiotherapy; however, most of the contemporary radiotherapy is carried out with linacs or teletherapy cobalt units.

5.2. X-RAY BEAMS AND X-RAY UNITS

Clinical x-ray beams typically range in energy between 10 kVp and 50 MV, and

are produced when electrons with kinetic energies between 10 keV and 50 MeV

are decelerated in special metallic targets.

In the target, most of the electron's kinetic energy is transformed into heat and a

small fraction of the energy is emitted in the form of x-ray photons which are

divided into two groups: characteristic x-rays and bremsstrahlung x-rays.

103

C hapter 5. Treatment Machines for External Beam Radiotherapy

5.2.1. Characteristic x-rays

Characteristic x-rays result from Coulomb interactions between the incident

electrons and atomic orbital electrons of the target material (collisional loss).

In a given Coulomb interaction between the incident electron and an orbital

electron, the orbital electron is ejected from its shell and the resulting orbital

vacancy is filled by an electron from a higher level shell. The energy difference

between the two shells may be emitted from the atom either in the form of a

characteristic photon (characteristic x-ray) or is transferred to an orbital electron

which is ejected from the atom as an Auger electron.

The fluorescent yield ω gives the number of fluorescent (characteristic) photons

emitted per vacancy in a shell (0≤ω≤1) and ranges from 0 for low Z atoms

through 0.5 for copper (Z = 29) to 0.96 for high Z atoms for K-shell vacancies that

are the most prominent sources of characteristic x-rays.

The photons emitted through electronic shell transitions have discrete energies

that are characteristic of the particular target atom in which the transitions have

occurred; hence the term characteristic radiation.

5.2.2. Bremsstrahlung (continuous) x-rays

Bremsstrahlung x-rays result from Coulomb interactions between the incident

electron and the nuclei of the target material.

During the Coulomb interaction between the incident electron and the nucleus, the

incident electron is decelerated and loses part of its kinetic energy in the form of

bremsstrahlung photons (radiative loss).

Photons with energies ranging from 0 to the kinetic energy of the incident electron

may be produced, resulting in a continuous bremsstrahlung spectrum.

The bremsstrahlung spectrum produced in a given x-ray target depends on the

kinetic energy of the incident electron as well as on the thickness and atomic

number Z of the target.

5.2.3. X-ray targets

In comparison with the range R of electrons of a given kinetic energy KE in the

target material, targets are divided into two main groups: thin and thick.

A thin target has a thickness much smaller than R, while the thickness of a thick

target is on the order of R.

For thin target radiation, the energy radiated is proportional to the product

(KE)×Z, where Z is the target atomic number. The intensity versus photon energy

(photon spectrum) is constant from 0 to KE, and 0 for all energies above KE.

104

Review of Radiation Oncology Physics: A Handbook for Teachers and Students

A thick target may be considered as consisting of a large number of superimposed thin targets. The intensity I(hν) of a thick target spectrum is expressed as: I(hν)=CZ(KE hν) , where C hν is a proportionality constant and is the photon energy. (5.1) X-rays are used in diagnostic radiology for diagnosis of disease and in radiation oncology (radiotherapy) for treatment of disease. X-rays produced by electrons with kinetic energies between 10 keV and 100 keV

are called superficial x-rays, with electron kinetic energies between 100 keV and

500 keV orthovoltage x-rays, and with electron kinetic energies above 1 MeV

megavoltage x-rays.

Superficial and orthovoltage x-rays are produced with x-ray tubes (machines),

while megavoltage x-rays are most commonly produced with linacs and sometimes with betatrons and microtrons.

Typical thin and thick target bremsstrahlung spectra originating from 100 keV

electrons striking a thin and thick target, respectively, are shown in Fig. 5.1.

内容需要下载文档才能查看

FIG. 5.1. Typical thin target (curve 1) and thick target (curves 2, 3, and 4) spectra for an x-ray tube in which 100 keV electrons strike the target. Curve (1) is for a thin target producing a constant intensity for photon energies from 0 to the kinetic energy of electrons striking the target (100 keV). Curve (2) represents unfiltered spectrum (inside the x-ray tube) for a thick target and represents a superposition of numerous thin target spectra; spectrum of curve (3) is for a beam filtered by x-ray tube window (low energy photons are filtered out); spectrum of curve (4) is for beam filtered by the x-ray tube window and additional filtration.

105

C hapter 5. Treatment Machines for External Beam Radiotherapy

5.2.4. Clinical x-ray beams

A typical spectrum of a clinical x-ray beam consists of line spectra that are

characteristic of the target material and are superimposed onto the continuous

bremsstrahlung spectrum.

The bremsstrahlung spectrum originates in the x-ray target, while the

characteristic line spectra originate in the target and in any attenuators placed into

the beam.

The relative proportion of the number of characteristic photons to bremsstrahlung

photons in an x-ray beam spectrum varies with electron beam kinetic energy and

atomic number of the target. For example, x-ray beams produced in a tungsten

target by 100 keV electrons contain about 20% characteristic photons and 80%

bremsstrahlung photons, while in the megavoltage range the contribution of

characteristic photons to the total spectrum is negligible.

In the diagnostic energy range (10 to 150 kV) most photons are produced at 90°

from the direction of electron acceleration, while in the megavoltage energy range

(1 to 50 MV) most photons are produced in the direction of electron acceleration

(forward direction: 0°).

5.2.5. X-ray beam quality specifiers

Various parameters, such as photon spectrum, half-value layer, nominal accelerating potential, beam penetration into tissue-equivalent media, etc., are used as x-ray beam quality indices(ee Sections 9.8.1 and 9.8.2 for details):

Complete x-ray spectrum is very difficult to measure; however, it gives the most

rigorous description of beam quality.

Half-value layer (HVL) is practical for beam quality description in the superficial

(HVL in aluminum) and orthovoltage (HVL in copper) x-ray energy range, but not

practical in the megavoltage energy range because in this energy range the

attenuation coefficient is only a slowly varying function of beam energy .

The effective energy of a heterogeneous x-ray beam is defined as that energy of a

monoenergetic photon beam that yields the same HVL as does the heterogeneous

beam.

Nominal accelerating potential (NAP) is sometimes used for describing the

megavoltage beam quality. The NAP is determined by measuring the ionisation

ratio in a water phantom at depths of 10 and 20 cm for a 10×10 cm2 field at the

nominal source-axis distance of 100 cm.

Recent dosimetry protocols recommend the use of tissue-phantom ratios or

percentage depth doses at a depth of 10 cm in a water phantom as an indicator of

megavoltage beam effective energy (beam quality index).

106

Review of Radiation Oncology Physics: A Handbook for Teachers and Students

5.2.6.

X-ray machines for radiotherapy Superficial and orthovoltage x-rays used in radiation therapy are produced with x-

ray machines. The main components of a radiotherapeutic x-ray machine are: an

x-ray tube; ceiling or floor mount for the x-ray tube; target cooling system;

control console; and an x-ray power generator. A schematic diagram of a typical

therapy x-ray tube is shown in Fig. 5.2.

The electrons producing the x-ray beams in the x-ray tube (Coolidge tube) originate in the heated filament (cathode) and are accelerated in vacuum toward

the target (anode) by an essentially constant-potential electrostatic field supplied

by the x-ray generator.

The efficiency for x-ray production in the superficial and orthovoltage energy

range is on the order of 1% or less. Most of the electron kinetic energy deposited

in the x-ray target (≈99%) is transformed into heat and must be dissipated

through an efficient target cooling system.

To maximize the x-ray yield in the superficial and orthovoltage energy range the

target material should have a high atomic number Z and a high melting point.

With x-ray tubes, the patient dose is delivered using a timer and the treatment

time must incorporate the shutter correction time (see Section 6.16) that accounts

for the time required for the power supply components to attain the steady state

operating conditions.

The x-ray tube current is controlled by hot filament emission of electrons which,

in turn, is controlled by the filament temperature (thermionic emission). For a

given filament temperature the x-ray tube current increases with the tube (anode)

voltage, first rising linearly with voltage in the space charge limited region and

saturating at higher voltages when all electrons emitted from the cathode are

pulled to the anode.

Research is currently carried out on cold field emission cathodes produced with

carbon nanotubes (CNT). The CNT-based cold cathode x-ray technology may

lead to more durable as well as miniature and portable x-ray sources for industrial

and medical applications.

内容需要下载文档才能查看

FIG. 5.2. Schematic diagram of a typical therapy x-ray tube (Reprinted from Johns, H.E. and

Cunningham, J.R. with permission).

107

版权声明:此文档由查字典文档网用户提供,如用于商业用途请与作者联系,查字典文档网保持最终解释权!

下载文档

热门试卷

2016年四川省内江市中考化学试卷
广西钦州市高新区2017届高三11月月考政治试卷
浙江省湖州市2016-2017学年高一上学期期中考试政治试卷
浙江省湖州市2016-2017学年高二上学期期中考试政治试卷
辽宁省铁岭市协作体2017届高三上学期第三次联考政治试卷
广西钦州市钦州港区2016-2017学年高二11月月考政治试卷
广西钦州市钦州港区2017届高三11月月考政治试卷
广西钦州市钦州港区2016-2017学年高一11月月考政治试卷
广西钦州市高新区2016-2017学年高二11月月考政治试卷
广西钦州市高新区2016-2017学年高一11月月考政治试卷
山东省滨州市三校2017届第一学期阶段测试初三英语试题
四川省成都七中2017届高三一诊模拟考试文科综合试卷
2017届普通高等学校招生全国统一考试模拟试题(附答案)
重庆市永川中学高2017级上期12月月考语文试题
江西宜春三中2017届高三第一学期第二次月考文科综合试题
内蒙古赤峰二中2017届高三上学期第三次月考英语试题
2017年六年级(上)数学期末考试卷
2017人教版小学英语三年级上期末笔试题
江苏省常州西藏民族中学2016-2017学年九年级思想品德第一学期第二次阶段测试试卷
重庆市九龙坡区七校2016-2017学年上期八年级素质测查(二)语文学科试题卷
江苏省无锡市钱桥中学2016年12月八年级语文阶段性测试卷
江苏省无锡市钱桥中学2016-2017学年七年级英语12月阶段检测试卷
山东省邹城市第八中学2016-2017学年八年级12月物理第4章试题(无答案)
【人教版】河北省2015-2016学年度九年级上期末语文试题卷(附答案)
四川省简阳市阳安中学2016年12月高二月考英语试卷
四川省成都龙泉中学高三上学期2016年12月月考试题文科综合能力测试
安徽省滁州中学2016—2017学年度第一学期12月月考​高三英语试卷
山东省武城县第二中学2016.12高一年级上学期第二次月考历史试题(必修一第四、五单元)
福建省四地六校联考2016-2017学年上学期第三次月考高三化学试卷
甘肃省武威第二十三中学2016—2017学年度八年级第一学期12月月考生物试卷

网友关注

音乐学科《小鸟,小鸟》教学过程
《角平分线性质》教学设计
《少年当自强》教学设计
小学品德《环球旅行》教案
教师资格面试小学语文《白杨》试讲答辩【语音示范】
小学品德与社会《五十六个民族五十六朵花》试讲答辩【语音示范】
《窃读记》教学设计
《胜利渡长江解放全中国》教学设计
小学音乐《唱山歌》试讲答辩【语音示范】
小学语文《爬天都峰》教学设计
以let's try和let'talk为例,展示小学英语听说课的设计思路
《木版画》教学设计
教资面试小学语文试讲逐字稿(《草原》第一课时)
小学数学的新课导入
小学数学《圆柱的表面积》试讲答辩【语音示范】
小学语文《去年的树》教学设计
小学音乐《飞跃彩虹》教学设计
音乐欣赏课教学设计《嘎达梅林》
小学信息技术《打字速度靠指法》教学设计
《吊饰》教学设计
《中华人民共和国国歌》教学设计
小学英语《Here is a red hat》试讲答辩【语音示范】
小学音乐《都有一颗红亮的心》教学设计
《五十六个民族五十六朵花》教学设计
《脚内侧停地滚球(水平三)》教学设计
小学英语《词汇-what can you do》试讲答辩【语音示范】
小学语文《山中访友》教学设计
教师资格证面试试讲如何才能让你出彩?
小学英语《Here is a red hat》教学设计
小学音乐《牧歌》教学设计

网友关注视频

沪教版八年级下册数学练习册21.4(1)无理方程P18
二年级下册数学第二课
二次函数求实际问题中的最值_第一课时(特等奖)(冀教版九年级下册)_T144339
沪教版牛津小学英语(深圳用) 四年级下册 Unit 12
河南省名校课堂七年级下册英语第一课(2020年2月10日)
【部编】人教版语文七年级下册《过松源晨炊漆公店(其五)》优质课教学视频+PPT课件+教案,江苏省
人教版二年级下册数学
19 爱护鸟类_第一课时(二等奖)(桂美版二年级下册)_T502436
8.对剪花样_第一课时(二等奖)(冀美版二年级上册)_T515402
苏科版数学八年级下册9.2《中心对称和中心对称图形》
七年级下册外研版英语M8U2reading
【部编】人教版语文七年级下册《逢入京使》优质课教学视频+PPT课件+教案,安徽省
苏科版数学 八年级下册 第八章第二节 可能性的大小
第19课 我喜欢的鸟_第一课时(二等奖)(人美杨永善版二年级下册)_T644386
第12章 圆锥曲线_12.7 抛物线的标准方程_第一课时(特等奖)(沪教版高二下册)_T274713
北师大版小学数学四年级下册第15课小数乘小数一
苏科版八年级数学下册7.2《统计图的选用》
沪教版八年级下册数学练习册一次函数复习题B组(P11)
沪教版牛津小学英语(深圳用) 四年级下册 Unit 8
沪教版牛津小学英语(深圳用) 五年级下册 Unit 10
化学九年级下册全册同步 人教版 第18集 常见的酸和碱(二)
冀教版小学数学二年级下册1
沪教版牛津小学英语(深圳用) 四年级下册 Unit 3
外研版英语三起5年级下册(14版)Module3 Unit2
冀教版英语四年级下册第二课
六年级英语下册上海牛津版教材讲解 U1单词
外研版英语七年级下册module3 unit1第二课时
【部编】人教版语文七年级下册《逢入京使》优质课教学视频+PPT课件+教案,辽宁省
二年级下册数学第一课
冀教版小学数学二年级下册第二单元《有余数除法的简单应用》